Is the body-protected area test every 6 months or every 12 months?
Every 12 months at most. AS/NZS 3003:2018 requires routine inspection and testing at intervals not exceeding 12 months, which is why the test label on the sign carries a date and the room is overdue once that date is a year old. Some testing companies describe it as “every 6 to 12 months”: 6 months is their offer, 12 months is the limit.
There is a 6-monthly step in some places, but it is not the same test. New Zealand’s RNZCGP Foundation Standard asks practices to press the test button on their RCDs every 6 months and record the date, who did it and the result, alongside the annual inspection by a registered electrical inspector. Some Australian contractors suggest the same push-button check. It is a useful habit that catches a dead RCD early, but it does not replace the 12-month test, which measures trip current and trip time with an instrument.
Alterations have their own rule: they go ahead only if the room has had its routine test in the previous 12 months.
Related
Rules in your state
Pick your state
See who certifies and inspects there.
More questions
- What does a new dental surgery need electrically?
- How many power points can a dental surgery have?
- Do I need an electrician to move a dental chair?
- How often does a medical or dental room need electrical testing?
- What is the green sign in a dental surgery or treatment room?
- Can a GP consult room be used for ECGs?
Need an electrician who does this?
Other states and New Zealand
Pick your state for its rules and the electricians listed there, or ask us to find one. Electricians and builders can get listed.
LND Group is the site publisher’s own business. How listings work.