Common questions
Can a portable 10 mA RCD replace a body-protected area?
Not as a substitute for wiring the room. A body-protected area is a property of the installation: the 10 mA RCDs in the room, one circuit per room, the identified outlets, the earthing and the sign, tested every 12 months. A plug-in or in-line 10 mA RCD protects whatever is plugged into it and nothing else, has to be switched on every time, and protects nobody if staff bypass it.
New Zealand guidance gives portable RCDs a limited place. The RNZCGP and NZIHE describe three situations where mains-powered equipment may be used on a patient outside a body-protected area: an emergency, such as an ECG after a collapse; a planned low-risk treatment with the device on a 10 mA in-line RCD while the practice works through a transition plan to a proper area; and low-risk appliances assessed as such. Equipment whose label shows a BF or CF applied part may not need the 10 mA RCD at all. Each of those is a stop-gap or an exception, documented in the practice’s policy, not a design.
In Australia the question is settled by classification: if the room is a patient area, it is wired as one. If you are using in-line RCDs today, record it, plan the upgrade, and test the RCDs you have.
Related
Rules in your state
Pick your state
See who certifies and inspects there.
More questions
- What does a new dental surgery need electrically?
- How many power points can a dental surgery have?
- Do I need an electrician to move a dental chair?
- How often does a medical or dental room need electrical testing?
- What is the green sign in a dental surgery or treatment room?
- Can a GP consult room be used for ECGs?
Need an electrician who does this?
Other states and New Zealand
Pick your state for its rules and the electricians listed there, or ask us to find one. Electricians and builders can get listed.
LND Group is the site publisher’s own business. How listings work.