Battery, plug-in and laptop ECGs: which need a body-protected area?

It turns on one thing: whether the patient is connected to something plugged into the wall while the leads are on. New Zealand’s RNZCGP scenarios (June 2026) work through the common cases.

  • A device that only runs on batteries and cannot be plugged in: no body-protected area needed.
  • A battery device that can also be plugged in, as many ECGs can: a body-protected area is preferred. Until there is one, the minimum is a 10 mA Type 1 medical-grade RCD and a sign on the device such as “Unplug me before connecting to the patient”.
  • A mains-only ECG, even plugged in through a 10 mA in-line medical-grade RCD: a certified body-protected area is needed. The in-line RCD does not replace it.
  • Any of these on an electric examination bed in a treatment room: a body-protected area is needed.
  • Off site, at an outreach clinic or marae: a risk-based approach, showing how electrical risk is reduced, what equipment is used (battery or mains) and what advice was obtained.

An ECG that runs from a laptop follows the same test: if the laptop is on its charger while the leads are on the patient, the patient is connected to mains-powered equipment. In Australia the same principle works through the room’s classification: the moment mains-powered equipment is applied to the patient there, the room is a patient area.

Related

Not quite your question? Ask it and we may answer it as a page.

Need an electrician who does this?

VictoriaLND Group, MelbourneMedical and commercial electricians. Installation, with the M class inspection and 12-month testing organised.Visit LND Group →REC 30162

Other states and New Zealand

Pick your state for its rules and the electricians listed there, or ask us to find one. Electricians and builders can get listed.

LND Group is the site publisher’s own business. How listings work.

Was this page useful?